How good or bad is the 10811 and which one on ebay is a better choice ?
I am looking for a very goo 5 MHz crystal oscillator with documentation .
Thanks, Ulrich N1UL
In a message dated 8/8/2015 7:16:37 P.M. Eastern Daylight Time,
***@n1k.org writes:
Hi
Ok, so John’s observation was the correct one. The data we are looking at
is *not* the
performance of the OCXO’s but the strange behavior of the counter at short
ADEV Tau’s.
Sorry for my bashing your poor 10811.
Bob
Post by t***@timeok.itHi all,
Luciano, how was the blue trace taken? Is this from your DMTD project?
If so, it's looking promising.
Post by t***@timeok.itThe green and magenta traces are definitely in the right ballpark for
measurements on a 5370-class counter. At 3E-11 @ t=12s the magenta trace is
optimistic but not outrageously so, while the green trace looks exactly like
I'd expect for a typical 10811 measured on a 5370. The observed noise is
due entirely to the counter until about t=200s. We see a glimpse of the
10811's typical ADEV at about 250 seconds, just before either drift or ADEV
uncertainty causes the trace to turn upwards. A longer run would be needed to
distinguish between these two situations.
Post by t***@timeok.it-- john, KE5FX
Miles Design LLC
r: no, all the measurements are taken using an HP53132A in frequency
mode. The difference can be the gate time. Using 1 second the resolution is
lower than using 2 Second that permit the max counter resolution.
Post by t***@timeok.itHi
Well an un-stated assumption of mine was that they all came from the
same measurement system and that it
Post by t***@timeok.ithad the same floor under all circumstances….
Bob
r: all the measurements are under the same conditions except for the
gate time of the counter.
Post by t***@timeok.itHi Luciano,
Can you give me the link to your ADEV posting image about 10811 vs 105
oscillators? I had it and now can't seem to find it. I wanted to look at
your plots as I read John Miles' comments. Many thanks.
Post by t***@timeok.itI have two very high performing HP10811-60109 OCXO units which I got
from Corby.
Post by t***@timeok.itMany thanks.
Jim Robbins
N1JR
r: Jim, I will load soon some files on my site. The 105B I have is a
fantastic exception, unfortunately it have a defect, may be a bad solder
inside cause randomly a phase jump and return to the original phase trend, so
all The long term ADEV are distorted by this problem I have to fix.
Post by t***@timeok.itTom VB, i will do all the cross measurements on 10811, 00105 and
rubidium I have and i will upload they but I need time to do this.
Post by t***@timeok.itUnfortunately I have only the HP53132A as TIC and my best reference are
four HP5065A. I normally use the counter function because the TI function
on 1 PPS have 100 time less resolution.
Post by t***@timeok.itWill be interesting doing also the Phase noise tests. I will do it.
http://www.timeok.it/files/time_and_frequency_house_standard_201r.pdf
Post by t***@timeok.itLuciano
www.timeok.it
Hi
Well an un-stated assumption of mine was that they all came from the
same
Post by t***@timeok.itmeasurement system and that it
had the same floor under all circumstances….
Bob
Hi
If that data is correct, then the 10811 you have is defective.
Bob
Well... some of the data is reasonable for a scenario where a counter
is
Post by t***@timeok.itbeing used to measure OCXOs.
Looking at the ADEV plot, I'd say the blue trace (HP105B vs 5065A) is
the most questionable one if it came from a standalone TIC or frequency
most circumstances. A Wavecrest box can measure at that level if it's
set
Post by t***@timeok.itup _perfectly_ to take bursts of 100+ wrap-free averages within a small
fraction of the t0 interval. It might also be doable with an HP 5370A/B
under similar conditions, but I'd have less confidence that the
averaging
Post by t***@timeok.itisn't distorting the measurement. So while It looks like a valid
measurement of an OCXO with some minor crosstalk or other external
interference, that may just be a coincidence.
Luciano, how was the blue trace taken? Is this from your DMTD project?
If so, it's looking promising.
The green and magenta traces are definitely in the right ballpark for
trace is
Post by t***@timeok.itoptimistic but not outrageously so, while the green trace looks exactly
like I'd expect for a typical 10811 measured on a 5370. The observed
noise
Post by t***@timeok.itis due entirely to the counter until about t=200s. We see a glimpse of
the
Post by t***@timeok.it10811's typical ADEV at about 250 seconds, just before either drift or
ADEV
Post by t***@timeok.ituncertainty causes the trace to turn upwards. A longer run would be
needed
Post by t***@timeok.itto distinguish between these two situations.
-- john, KE5FX
Miles Design LLC
_______________________________________________
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1]
and follow the instructions there.
_______________________________________________
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [2]
and follow the instructions there.
------
[1]
http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
Post by t***@timeok.itilman/listinfo/time-nuts[2]
http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
Post by t***@timeok.itilman/listinfo/time-nuts
Message sent via Atmail Open - http://atmail.org/
_______________________________________________
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
Post by t***@timeok.itand follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-***@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-***@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.